What three test loads are used to calibrate an RF vector network analyzer?

  1. 50 ohms, 75 ohms, and 90 ohms
  2. Short circuit, open circuit, and 50 ohmsBonne réponse
  3. Short circuit, open circuit, and resonant circuit
  4. 50 ohms through 1/8 wavelength, 1/4 wavelength, and 1/2 wavelength of coaxial cable

Explication

Calibration solves for the analyzer's systematic error terms by measuring standards whose reflection is known and widely separated, namely a short, an open and a precision load equal to the system impedance. Assorted fixed resistors or lengths of cable do not provide that set of known reflections.